Full Version: (32S) Determination of Dry Matter in small silage samples
An excerpt from Determination of Dry Matter in Small Silage Samples, Journal of the University of Wales Agricultural Society, 78, 101-114 (1988)
        INTRODUCTION AND METHODOLOGY
A previous paper (Faithfull, 1997) indicated that a suitable method for dry
matter determination in small silage cores had been developed; this paper
describes that method. …
        Calculation
… the volume of aqueous distillate is reduced by the volume of acid. This is estimated
according to the formula …
        Programmable calculator program
A program has been devised to allow easy calculation using a Hewlett-Packard
programmable calculator; this is shown in Figure 3.
        Figure 3. Silage Dry Matter Program for Hewlett-Packard HP 32SII
To write the program into the calculator, activate program-entry mode and start
at the top by typing: …
To run the program, enter the variables and store them. The variables are as
follows:
BEST!
SlideRule
I really appreciate your posts of the use of HP calculators in published papers.
In this case however I noticed a few issues:
  • useless use of ENTER
  • division of the constant 0.00555 by 10
  • useless use of register B
  • use of constant 10 instead of register M
While the first three points are just a bit of annoyance, the last one on line 017 is really a bug.
You might enter a value different from 10 for \(m\) in register M and it will not be taken into consideration.
Here's a corrected and slightly optimised version of that program:
LBL W
1
RCL T
5.55E-4
RCL V
998
RCL M
1E3
x<>y
RTN
Example
10 STO M
7.5 STO V
10 STO T
XEQ W
255.6542
References
Thank you for the kind note & the technical review; much appreciated by all!
BEST!
SlideRule