Difference between Real and Virtual Calculator
10-04-2017, 11:30 AM
Post: #3
RE: Difference between Real and Virtual Calculator
The program is working on my real calculator with no problem.
the purpose of this program is to check and calibrate thermocouples.
the thermocouple is a sensor that is giving a resistance (in Ohm) changing with the temperature. PT100 is a thermocouple that give 100 Ohm at 0°C
PT1000 is a thermocouple that give 1000 Ohm at 0°C
They follow the equation:

R=D*(1+A*T+B*T^2+C*(T-100)*T^3)

where T is the temperature, D is the 0°C-resistance (100 or 1000) and the remaining parameters A, B and C are:
A = 3.9083E10^-3
B = -5.7750E10^-7
C = -4.1830E10=12 for t<0 or 0.0 for t>=0


Syntax:
RTD(r,t,d)
r= Reststance
t= Temperature
d=Thermocouple Type. the most common values are 100 or 1000


Example:
RTD(0,26.5,100) will give 110.397549938, that is the Resistance value for 26.5°C for a PT100 sensor

RTD(124.5,0,100) will give 62.1074171403, that is the Temperature value for 124.5 Ohm for a PT100 sensor